    Why might an agency want an assay validation range to be wider than a given specification? (agency commented that validated range too narrow).


  • 2.  RE: Assay validation range versus spec

    Posted 29-Nov-2022 09:36
    Anon, a quick reference on this would be ICH Q2(R1) Section II.5 on Range. Generally speaking, you'll need confidence in how far outside of a given spec the result may be, to a reasonable extent.

    ------------------------------
    Jonathan Amaya-Hodges
    Director, Technical Services
    Sharon MA
    United States
    ------------------------------



  • 3.  RE: Assay validation range versus spec

    This message was posted by a user wishing to remain anonymous
    Posted 29-Nov-2022 11:05
    This message was posted by a user wishing to remain anonymous

    In short, because you have to have as much confidence in the accuracy of results that are at the limit of specification as within specification.  I recommend you follow ICH Q2 (R1) - it is a thorough description of the minimum data required to validate an analytical method and is widely accepted by global Agencies.  Validation of assay range is usually 80 - 120 %.


  • 4.  RE: Assay validation range versus spec

    Posted 29-Nov-2022 13:46
    The validated range needs to allow for OOS (out of spec) findings. As a result, the validated range must be able to accurately test not only the range intended by the specification chosen, but the validated range also needs to allow/test for "real" values falling either HIGHER or LOWER than the range of values intended). In other words, the "actual" value reported needs to be VALID even if OOS.

    ------------------------------