Qiagen Sciences N.V. - ENTERPRISE
Original Message:
Sent: 01-Apr-2024 11:53
From: Richard Vincins
Subject: IVDs & MDs discontinuation
Good day Oleksii,
Since you are speaking about General IVD and List B, then no there would not need to be a notification for discontinuation of a product to the Notified Body. But I completely agree with Christoph you may have to notify them anyway based on contractual arrangement. If you have a List B product that is being discontinued, you would need to notify them that those product(s) would no longer be under scope of the certificate or you are withdrawing the Declaration of Conformity - thus the certificate would "stop" also. In addition, there should be an "end of life" plan established for these products, because depending on the shelf-life/expiration date, the Notified Body may still be reviewing information, so helpful for the notification and when things with the product "stop" completely.
Since there is no Notified Body intervention under the EU IVDD for General List IVD, there is no obligation to a Notified Body. Though again, if your quality system certificate encompasses certain processes or activities, this may need notification as well. It is frustrating for an auditor to show up at the facility ready to do their audit, start asking questions about Product A, and the company says, "Oh yea, by the way, we stopped selling that product 8 months ago. It is not in scope anymore." This could have a significant impact on the audit agenda or planning of the auditor.
------------------------------
Richard Vincins ASQ-CQA, MTOPRA, RAC
Principal Strategy Consultant
NAMSA
Original Message:
Sent: 29-Mar-2024 15:25
From: Anonymous Member
Subject: IVDs & MDs discontinuation
This message was posted by a user wishing to remain anonymous
Hi Oleksii
Per the IVDD, 98/79/EC, Article 10, any discontinuation has to be notified. Per the IVDR, 2017/746, Annex VI, EUDAMED will be updated when a product is no longer available on the market. So I guess the NB expects to be informed.
I wonder if your NB contract mentions product discontinuations.
Original Message:
Sent: 26-Mar-2024 09:28
From: Oleksii Belenytskyi
Subject: IVDs & MDs discontinuation
Hello All,
A company is not going to distribute several devices. We got a corresponding change request. General IVDs and Annex II, List B devices affected by this change. We use NBOG 2014-3 and MDCG 2022-6 to assess changes. I would appreciate any ideas about how discontinuation changes should be assessed. For example, we came up with several options:
1) General IVDs – non-significant change, notification to NB is not necessary
Annex II, List B – significant, NB is to be notified
2) General IVDs – significant change, notification to NB is not necessary
Annex II, List B – significant, NB is to be notified
3) General IVDs – non-significant change, notification to NB is not necessary
Annex II, List B – non-significant (as not impact on the safety and effectiveness), NB is to be notified
Please share your opinion on which option you support and why?
Or, please suggest another option.
Regards,
Oleksii
------------------------------
Oleksii Belenytskyi
Qiagen Sciences N.V. - ENTERPRISE
Wrocław
Poland
------------------------------