In my experience in more than 25 years of being audited by multiple of the most prolific and difficult European Notified Bodies (e.g., TUV, BSi, and others), I've never had one demand an outside test house/lab report showing my or my clients' conformity with EN ISO 14971 as amended (nor with the old EN 1441 that we used to use before EN ISO 14971 was born). Instead, it has always been sufficient to demonstrate conformity via appropriate objective evidence.
Indeed, the main risk management requirements of the EU MDR [i.e., Articles 10(2), 10(4), 10(9) third paragraph indent (e), Annex I.3, Annex II.5, etc.] don't demand an official report from an outside test house/lab. Instead, they require, "technical documentation" that "contain information on...the solutions adopted..." [emphasis added], and, "technical documentation" that is "...such as to allow the conformity of the device with the requirements of [the EU MDR] to be assessed. [emphasis added].
Your in-house objective evidence of conformity could certainly take the form of a "test report", but could also be another type of documented evidence such as itemization in your GSPR checklist, a memo, a separate trace matrix, etc.
------------------------------
Kevin Randall, ASQ CQA, RAC (Europe, U.S., Canada)
Principal Consultant
Ridgway, CO
United States
© Copyright 2022 by ComplianceAcuity, Inc. All rights reserved.
------------------------------
Original Message:
Sent: 25-Oct-2022 10:07
From: Anonymous Member
Subject: When you perform a gap analysis and find you meet the updated standard, do you update your DOC and IFU with the latest?
This message was posted by a user wishing to remain anonymous
If you have a test report stating that you meet ISO 14971:2007, but you perform an internal gap analysis and update your documentation to meet ISO 14971:2019, can you say that you meet ISO 14971:2019 in your documentation, e.g., in your IFU and in your EU DoCs? Or are you required to have an official test report stating you meet the updated standard?