Also, depending on how the laptop and the mouse interface with the medical device but likely you will also need to consider IEC 60601-1, clause 16 requirements (Medical Electrical System Requirements). You may want to refer to Annex I of IEC 60601-1 as it will show you different configurations that may apply and how to isolate properly. Yes, Usability per IEC 60601-1-6 & 62366 (matching to 60601-1-6 but for FDA and many other regulators would be IEC 62366-1), you may have life cycle software to deal with as well under IEC 62304. Lastly, you should also need to consider IEC 60601-1-2 (EMC). Other standards may apply such as IEC 60601-1-11 for home use medical electrical devices and IEC 60601-1-12 emergency medical services medical electrical equipment and a whole host of Particular standards under the IEC 60601-2-XX or IEC/ISO 80601-2-XX.
So, you will likely have verification testing and validation testing involved in this type of design you have very briefly described.
------------------------------
Leonard (Leo) Eisner, P.E.
The "IEC 60601 Guy"
Principal Consultant, Eisner Safety Consultants
Phone: (503) 244-6151
Mobile: (503) 709-8328
Email:
Leo@EisnerSafety.comWebsite:
www.EisnerSafety.com------------------------------
Original Message:
Sent: 21-Sep-2019 09:42
From: Anonymous Member
Subject: Laptops to be tested as medical devices
This message was posted by a user wishing to remain anonymous
Hi,
If a laptop and wireless mouse is to be used with a medical device , does it need to be tested as part of the medical device testing . If the manufacturer doesn't provide these items with the medical device and claims in the IFU on use of any laptop with the minimum requirements , is that acceptable?