  • 1.  Animal Study and Usability

    This message was posted by a user wishing to remain anonymous
    Posted 10-Mar-2023 08:28
    This message was posted by a user wishing to remain anonymous

    Is it acceptable/common to collect usability data during an Animal Study? 

    We have a product very similar to the predicate. 

    We would like to justify HFcategory 2  (FDA)  can we use the animal study to collect the data?  



  • 2.  RE: Animal Study and Usability

    Posted 10-Mar-2023 10:26

    As a general rule, my understanding is that the HF Category is intrinsically governed by the nature of the subject device and its intended use rather than the HF study method or results.  For example, HF Category 2 devices are expected by FDA to involve no user tasks which, if performed incorrectly or not performed at all, would or could cause serious harm to the patient or user, where harm is defined to include compromised medical care.

    In general for assessing the usability of a subject device, there could be a variety of user interface test-usage paradigms (e.g., use on humans, use on animals, or even on the bench) that could be sufficient as long as the chosen paradigm is representative of how the user interacts with the user interface.  For example, testing a user's interaction with an electrosurgical generator and handpiece by having the user use the devices on animal tissue could be an adequate way to assess the usability as long as the testing gives appropriate consideration to relevant human factors like user characteristics, capabilities, needs, expectations, cognition, preferences, real-world demands, etc., etc.



    ------------------------------
    Kevin Randall, ASQ CQA, RAC (Canada, U.S., Europe)
    Principal Consultant
    Ridgway, CO
    United States
    © Copyright 2023 by ComplianceAcuity, Inc. All rights reserved.
    ------------------------------



  • 3.  RE: Animal Study and Usability

    Posted 10-Mar-2023 10:39

    Hi ,

    Category 2 of the FDA's Human Factors guidance refers to medical devices that have moderate-to-high user interface risk. This means that the device's user interface could pose a potential risk to the safety and effectiveness of the device if not properly designed, tested, and validated. Examples of Category 2 medical devices include infusion pumps, patient monitors, and surgical instruments.

    Yes, the Human Factors (HF) Category 2 designation can be used to guide the usability study of medical devices with moderate-to-high user interface risk. In fact, usability testing is one of the key activities recommended by the FDA's HF guidance for Category 2 medical devices, as it provides critical insights into how users interact with the device and helps identify potential use-related hazards. The results of the usability testing should be documented in a Human Factors Engineering/Usability Engineering Report.



    ------------------------------
    Raje Devanathan
    Amerisource Bergen
    TPIreg, Innomar Strategies
    Senior Manager - Regulatory Affairs, Medical Devices
    rdevanathan@tpireg.com
    3470 Superior Court
    Oakville ON L6L0C4
    Canada
    ------------------------------



  • 4.  RE: Animal Study and Usability

    Posted 10-Mar-2023 10:53

    Raje,

    Since FDA says that Category 2 devices require no "critical" tasks (i.e., they involve no user tasks which, if performed incorrectly or not performed at all, would or could cause serious harm to the patient or user, where harm is defined to include compromised medical care), could you give more explanation about your assertion that Category 2 devices instead have "moderate-to-high" user interface risk?



    ------------------------------
    Kevin Randall, ASQ CQA, RAC (U.S., Europe, Canada)
    Principal Consultant
    Ridgway, CO
    United States
    © Copyright 2023 by ComplianceAcuity, Inc. All rights reserved.
    ------------------------------



  • 5.  RE: Animal Study and Usability

    Posted 13-Mar-2023 15:22

    Hi Kevin, While it is true that Category 2 devices do not involve critical tasks, it does not mean that they do not have moderate-to-high user interface risks. Category 2 devices may require the user to interact with the device's user interface to perform tasks that, if performed incorrectly, could lead to adverse events or compromised medical care.

    For example, a Category 2 device could be a blood glucose meter that requires the user to enter the correct information, such as the date and time, before performing a glucose test. If the user enters incorrect information, it could result in inaccurate test results, leading to incorrect treatment decisions and compromised medical care.

    Another example of a Category 2 device is a medication dispensing system that requires the user to enter the correct dosage and medication information. If the user enters incorrect information, it could result in the wrong medication being dispensed, leading to adverse events and compromised medical care.

    Therefore, while Category 2 devices do not involve critical tasks, they still have moderate-to-high user interface risks that need to be carefully considered and mitigated to ensure patient safety.



    ------------------------------
    Raje Devanathan
    Amerisource Bergen
    TPIreg, Innomar Strategies
    Senior Manager - Regulatory Affairs, Medical Devices
    rdevanathan@tpireg.com
    3470 Superior Court
    Oakville ON L6L0C4
    Canada
    ------------------------------



  • 6.  RE: Animal Study and Usability

    Posted 13-Mar-2023 15:45

    Hi Raje, sounds like you're saying that blood glucose meters and medication dispensing systems don't involve critical tasks, i.e., that they don't involve user tasks which, if performed incorrectly or not performed at all, would or could cause serious harm to the patient or user, where harm is defined to include compromised medical care).  What would be the rationale for claiming that that blood glucose meters and medication dispensing systems don't involve user tasks which, if performed incorrectly or not performed at all, would or could cause serious harm to the patient or user, where harm is defined to include compromised medical care?



    ------------------------------
    Kevin Randall, ASQ CQA, RAC (U.S., Europe, Canada)