The EU MDR doesn't prescribe the frequency at which CERs are to be updated. Modeling this, the MDCG (2020-13) has left manufacturers the liberty to, "...Identify when updates to the clinical evaluation report shall be assessed during the surveillance and post certification monitoring activities and which frequency should be considered...". Consequently, yes, the frequency of updates is generally up to the manufacturer's discretion. Yet that must be appropriate risk-based discretion.
While MEDDEV 2.7/1 revision 4 gives specific guidelines (specifically, update at least annually if the device carries significant risks or is not yet well established; or every 2 to 5 years if the device is not expected to carry significant risks and is well established), we need to remain sensitive to the fact that MEDDEV 2.7/1 isn't generally intended for the EU MDR. Nonetheless, for now, the Notified Bodies and other Union authorities are still often generally tolerant of that MEDDEV in my experience as along as its prescription is reasonable for the subject device. Otherwise, the manufacturer is required to tailor the frequencies to its particular device scenario.
------------------------------
Kevin Randall, ASQ CQA, RAC (Europe, U.S., Canada)
Principal Consultant
Ridgway, CO
United States
© Copyright 2023 by ComplianceAcuity, Inc. All rights reserved.
------------------------------
Original Message:
Sent: 21-Feb-2023 11:12
From: Rajeswari Devanathan
Subject: Clinical Evaluation Report Frequency
Hi,
Clause 6.2.3 requires the CER to be updated at least annually for high risk or new devices, and every 2 to 5 years for lower risk, well-established devices. A justification for the frequency of updates will be required. For all risk classifications of devices, the CER will need to be updated whenever new information from the Post Market Surveillance (PMS) affects the evaluation or its conclusions.
------------------------------
Raje Devanathan
Amerisource Bergen
TPIreg, Innomar Strategies
Senior Manager - Regulatory Affairs, Medical Devices
rdevanathan@tpireg.com
3470 Superior Court
Oakville ON L6L0C4
Canada
Original Message:
Sent: 21-Feb-2023 10:16
From: Anonymous Member
Subject: Clinical Evaluation Report Frequency
This message was posted by a user wishing to remain anonymous
Under the EUMDR, is it up to the manufacturer to establish the frequency of which CERs are updated? Is it the same for all device classifications? Where is this defined? Is there flexibility?
Thanks in advance btw