Regulatory Open Forum

 View Only
  • 1.  Clinical performance study of Class A non-sterile product not required for IVDR?

    This message was posted by a user wishing to remain anonymous
    Posted 05-Apr-2023 13:35
    This message was posted by a user wishing to remain anonymous

    I'm working on Technical Documentation for transitioning a Class A non-sterile cytometer from IVDD to IVDR. The instrument itself does not generate any diagnostic result, it is intended for use with IVD reagents to identify lymphocyte subsets. The instrument is Class A non-sterile, therefore is over due for IVDR compliance. The reagent is Class C and still has about 3 years of transition time. I'm reusing all the previous IVDD data trying to get through IVDR without any additional testing.
    I've found on some consulting companies' website indicating that clinical performance study is not expected for Class A product, therefore clinical performance study report (CPSR) is not required. But I don't see this kind of statement in any MDCG or other guidelines. Would you please let me know if this is true? 
    Thank you very much!


      




  • 2.  RE: Clinical performance study of Class A non-sterile product not required for IVDR?

    Posted 06-Apr-2023 05:35

    Good day Anon,

    It will not be specifically stated a Class A, non-sterile classed device does not require a Clinical Performance Study or associated report.  That determine needs to be made by the organisation within the context of the Performance Evaluation, Annex XIII.  If reading through Annex XIII there is no clear distinction which devices or class of devices would require performance evaluation or not.  All IVD medical devices require performance evaluation, but the activities within performance evaluation are determined, justified, and rationale generated for why components of performance evaluation would not be required.  As an example specifically from Section 1.2.3 it states: 'Demonstration of the clinical performance The manufacturer shall demonstrate the clinical performance of the device in relation to all the parameters described in point (b) of Section 9.1. of Annex I, unless any omission can be justified as not applicable.'  Emphasised and you will find similar wording throughout the Annex.  There is also additional guidance in MDCG 2022-2 as example where scientific validity for Class A devices does not need to be done if can be justified.

    The answer being all IVD medical devices require performance evaluation.  It is up to the manufacturer to define, defend, and provide sufficient clinical evidence why the IVD medical device would not require further clinical evidence and/or clinical data to support the devices safety and performance.  Again, going back to Annex XIII there is wording about the scientific validity, analytical performance, and clinical performance which requires each or a combination of each.  Therefore, it is up to the Manufacturer to define and justify whether clinical performance study would be required or not.



    ------------------------------
    Richard Vincins ASQ-CQA, MTOPRA, RAC
    Vice President Global Regulatory Affairs
    ------------------------------



  • 3.  RE: Clinical performance study of Class A non-sterile product not required for IVDR?

    This message was posted by a user wishing to remain anonymous
    Posted 07-Apr-2023 08:02
    This message was posted by a user wishing to remain anonymous

    Hi Richard,

    Thank you very much for the quick reply, and appreciate the clarification, it helps.

    Hello everyone,

    Just wondering if there is any good examples of justification statements on the omission of clinical performance study for any IVD device. Thanks again.