Thank you Anne for pointing out that narrative, which nicely balances out the aforesaid impractical and nonsensical MDCG narrative. Indeed, since the scope of Q&A #11 (see its title) encompasses SPHTs and then states the general rule I recited before, then I think we will run into notified bodies or other regulators who might try to assert the timeline I recited before. If so, then I think your recitation will be a good rebuttal.
© Copyright by ComplianceAcuity, Inc. All rights reserved.
Original Message:
Sent: 20-Jan-2024 16:20
From: Anne LeBlanc
Subject: Complaint Handling and Reporting - MDR - Serious Public Health Threat
Hi Kevin
Yes, for serious incidents in general, and potentially serious incidents, awareness usually begins when the complaint is received. Specifically for serious public health threats, though, see MDCG 2023-3 sections
"A serious threat to public health will in principle not be limited to one isolated case or individual patient issue, and identifying these events may depend on signal detection or trending of multiple events of the same nature/typology, same root cause, etc." And
"A serious public health threat must be reported immediately, and not later than 2 days after the manufacturer becomes aware of that threat (Article 87(4) MDR)"
Then the report will probably go on to explain what pieces of information were received when and how the manufacturer came to that state of awareness.
Practically and sensibly, I'd hope when SPHTs do happen there's a channel available for immediate recognition and escalation, to solve the problem as well as reporting it.
------------------------------
Anne LeBlanc
United States
Original Message:
Sent: 20-Jan-2024 10:34
From: Kevin Randall
Subject: Complaint Handling and Reporting - MDR - Serious Public Health Threat
Notwithstanding considerations of practicality or sensibility, MDCG 2023-3 seems quite clear that we aren't allowed to wait until after our investigation to start the clock. Instead, MDCG 2023-3 starts the clock when the complaint is received. Here are a couple excerpts:
- "...The awareness date (day=0) refers to the date when the manufacturer is first made aware or receives information of the occurrence of the (potentially) serious incident and not after it has conducted its investigation..." [emphasis added]
- "...the 'manufacturer awareness date' is the date when the first employee or representative of the manufacturer's organisation, receives information (e.g. a complaint) regarding the potentially serious incident..." [emphasis added]
------------------------------
Kevin Randall, ASQ CQA, RAC (Europe, U.S., Canada)
Principal Consultant
Ridgway, CO
United States
© Copyright by ComplianceAcuity, Inc. All rights reserved.
Original Message:
Sent: 20-Jan-2024 10:17
From: Christina Donegan
Subject: Complaint Handling and Reporting - MDR - Serious Public Health Threat
Thanks for that Anne.
Best Regards,
Chris.
Original Message:
Sent: 1/19/2024 8:26:00 AM
From: Anne LeBlanc
Subject: RE: Complaint Handling and Reporting - MDR - Serious Public Health Threat
Hi Chris
I think you are right that in most cases the SPHT will be identified through investigation, so the clock will begin at the time when the investigation reaches that conclusion. Not necessarily when the investigation is complete, because that might take longer. And not necessarily when the complaint first comes in, because you might not know at first that there is a SPHT, although if there is a suspicion of something serious you will want to expedite the investigation.
Trying to imagine a case where a SPHT might be identified on receipt of a complaint... Maybe if you had a threshold for a number of incidents, one complaint would be the one that crossed the threshold. Maybe one complaint says something alarming like "we tested your sterile substance on receipt and found massive contamination".
Complaint handlers may not have the expertise to judge SPHT for any given complaint, but they probably do have the expertise to select a subset of complaints that warrant escalation for additional review.
------------------------------
Anne LeBlanc
United States
Original Message:
Sent: 15-Jan-2024 14:30
From: Christina Donegan
Subject: Complaint Handling and Reporting - MDR - Serious Public Health Threat
Thanks everyone for your feedback on this. My only concern is that I cannot envisage a scenario whereby a single complaint could or have the ability to determine whether a serious incident is a serious public health threat. I understand all the timelines, but my question is more related to how should the process for this particular category should work in terms of assessing for SPHT? Should it be determined and reported once investigation and escalation are completed? How would a complaint handler or vigilance SME be able to assess such a complaint without other expertise and review, e.g. a complaint is received that states that there is a serious public health threat due to a fault in a single product or batch of products?
I am just curious how others are handling this?
Thanks in advance,
Chris.
Original Message:
Sent: 1/5/2024 9:03:00 AM
From: Tom Patten
Subject: RE: Complaint Handling and Reporting - MDR - Serious Public Health Threat
Common misconception here is that it is always 2 days and it's a big fat can of worms but in the above example where the last day of a period is a bank holiday, saturday or sunday the reporting period shall end at the end of the following working day. So depending on your member state where the 25th and 26th are public holidays and you could have until the 27th to report.
I do full agree that best practice as ever - is to report immediately.
------------------------------
Tom Patten,
NSAI,
Ireland.
Original Message:
Sent: 05-Jan-2024 02:29
From: Ronald Boumans
Subject: Complaint Handling and Reporting - MDR - Serious Public Health Threat
Hi Christina,
As Bryndis Gudnadottir already mentioned, see MDCG 2023-3. This states that the clock starts ticking at midnight (your local time) at the end of the day you first became aware. These are callendar days. As an example, if you would have learned last year on Friday 22 December, you would have had until Sunday 24, Christmas evening, to submit.
If you want to have a smart procedure, just be aware that you don't need to label an incident as being a serious public health threat. It can be anything, it is up to the authorities to label this event. By reporting within two days, you will be in time regardless of the type of incident. It is also important to realise, you can enter a report that is still missing data and update later. With that in mind, I would recommend to develop a procedure that quickly identifies potential high-risk cases and escalates them to a high level of attention. Bottom line: when in doubt, report immediately and fill in the blanks later. If you do this, you will make life for the vigilance staff (and for the rest of the organization) much easier.
For those thinking they will draw attention if they report incidents, just realise that the authorities have dozens of incidents coming in every day. For them, these reports are routine matters. In the case you draw attention, it will most likely be in the form of scoring bonus points for being early with your reports. I know this, because I worked there and helped developing risk based reporting systems.
------------------------------
Ronald Boumans
MDR Expert
Super PRRC
Netherlands
Original Message:
Sent: 04-Jan-2024 02:09
From: Christina Donegan
Subject: Complaint Handling and Reporting - MDR - Serious Public Health Threat
Hi Folks,
In terms of Article 87 and the timelines for vigilance, I would like to ask how is the 2 day timeline for SPHT handled? Does the clock start once a single complaint is received or when a SPHT is determined via investigation. Notified Bodies are focusing on this in audits and I wanted to see what your thoughts were on this.
Thank You,
Chris.