The manufacturer, you, need to write the test plan. In IEC 60601-1-2:2014+A1:2020, sub-cl 6.2 requires that a test plan shall be provided to the test lab (i.e. the manufacturer needs to provide it to the test lab as the test lab can't determine how to test your product you know it better than anyone else).
Annex G provides Guidance on the recommended content of a test plan.
Please note one of the critical items of an EMC test plan that is missed and that the FDA is very critical about is that Essential Performance needs to well defined and clear in the test plan as that has a significant impact on the EMC testing.
------------------------------
Leonard (Leo) Eisner, P.E.
The "IEC 60601 Guy"
Principal Consultant, Eisner Safety Consultants
Mobile: (503) 709-8328
Email:
Leo@EisnerSafety.comWebsite:
www.EisnerSafety.com------------------------------
Original Message:
Sent: 15-Jan-2024 03:41
From: Anonymous Member
Subject: Documents needed for EMC testing
This message was posted by a user wishing to remain anonymous
Dear Colleagues,
We are preparing for EMC testing of our medical device (60601-1-2 and EMC requirements from some particular standards) for compliance in three different markets of interest.
The question is, what kind documentation do we need to prepare for the testing lab (e.g. UL)? Do we, the manufacturer, need to write a test plan? How is this responsibility delegated?
Any input will be a great guidance and advice. Thank you!