Anon,
Also for clarification, the IEC 60601-1-6 is applicable to Electromechanical medical device products only - whereas IEC 62366-1 is applicable to all medical devices. At one point they were actually going to remove IEC 60601-1-6 since IEC 62366 existed, but then much of the content was moved to IEC 62366-1. As Anne indicated, today the standard is really just a "bridge" between IEC 60601-1 usability/use requirements and IEC 62366-1.
If you have an Electromechanical device it would be proper to cite both IEC 60601-1-6 and IEC 62366-1. If your product is non-electromechanical, then citing of only IEC 62366-1 would be needed.
------------------------------
Richard Vincins ASQ-CQA, MTOPRA, RAC
Vice President Global Regulatory Affairs
------------------------------
Original Message:
Sent: 02-Aug-2023 19:35
From: Anne LeBlanc
Subject: Does IEC 62366-1 replace IEC 60601-1-6 ?
Not entirely. The 62366 standards were built upon the older 60601-1-6. Then 60601-1-6 was revised to strip out most of its content in favor of pointers to 62366-1 and 62366-2. There are just a few requirements remaining in 60601-1-6. The next revision of the general standard may absorb those last few bits of 60601-1-6 and point directly to 62366-1 as the collateral standard.
The introduction to 60601-1-6 describes its current status as a sort of middleman: https://webstore.iec.ch/preview/info_iec60601-1-6%7Bed3.2%7Db.pdf
------------------------------
Anne LeBlanc
United States
Original Message:
Sent: 02-Aug-2023 12:41
From: Anonymous Member
Subject: Does IEC 62366-1 replace IEC 60601-1-6 ?
This message was posted by a user wishing to remain anonymous
I am asking since IEC 60601-1 reference IEC 62366-1 and IEC 60601-1-6 and FDA recognizes both
Thanks,