  • 1.  EMDN Code (EU MDR Requirement)

    This message was posted by a user wishing to remain anonymous
    Posted 08-Sep-2022 08:15
    This message was posted by a user wishing to remain anonymous

    I would appreciate it if someone could explain the requirement to obtain the EMDN code per Article 26 of the MDR.
    1. Is that a must ?
    2. Apply to all devices? 
    3. does it have a transition time? 
    4. any info you can provide on the topic. 

    MTIA


  • 2.  RE: EMDN Code (EU MDR Requirement)

    Posted 08-Sep-2022 10:49
    Edited by Kevin Randall 08-Sep-2022 10:50

    From a pure legislative perspective, the EU MDR's Article 26 requirement for a medical device nomenclature code is currently only voluntary.  Nonetheless, you may find certain Notified Bodies contravening this by mandating nomenclature codes. So, it's important to know your Notified Body (if any is involved in your particular conformity assessment process).

    The nomenclature code requirement is only voluntary because a) it is tethered to the EUDAMED database device registration requirements of Articles 27-29 and 33, and because b) currently, pursuant to Articles 34 and 123, those EUDAMED device registration requirements are only voluntary until such time as the entire EUDAMED system has been declared fully functional following an independent audit and corresponding Commission notice to be published in the Official Journal.

    The EUDAMED database consists of six modules.  Only three are currently available for voluntary use.  One of those is the UDI/device registration module which, pursuant to Annex VI.B.8, is where the nomenclature code is to be entered.

    Pursuant to Article 29, the assignment of the nomenclature code won't generally apply to custom-made devices.  Investigational devices may also be exempt from the nomenclature code requirement since they are exempt from the Article 27 UDI requirement, and thus the Annex VI.B.8 nomenclature code entry. I know of no other device-specific exemptions from the nomenclature code requirement.




    ------------------------------
    Kevin Randall, ASQ CQA, RAC (Europe, U.S., Canada)
    Principal Consultant
    Ridgway, CO
    United States
    © Copyright 2022 by ComplianceAcuity, Inc. All rights reserved.
    ------------------------------



  • 3.  RE: EMDN Code (EU MDR Requirement)

    Posted 09-Sep-2022 04:39

    Hi Anon

    There are a couple of areas of MDR (&IVDR) where the EMDN code is referenced: for Registration of devices (needed as part of assignment of UDI-DI, Annex VI part B; from Art. 29 & 33; as mentioned already) and they are referenced under the MDCG Guidance on Grouping & Sampling (MDCG 2019-13) for grouping of devices. The latter guidance comes from the requirement under Article 52 regarding conformity assessment, and the requirement for assessment of technical documentation of 'at least one representative device' per group.

    Therefore, the device codes are used by the Notified Bodies at the point of Application to support grouping of devices appropriately (in order to then plan an appropriate conformity assessment; see requirement Annex IX 2.1 2nd sub-bullet), and in the case of devices that require a product specific certificate, that these codes are assigned appropriately; linking to the requirement for Basic UDI-DI on these certificates.

    Therefore, Yes, EMDN codes are needed, as you need them at the time of Application to a NB; the NB has to show that they have met all the requirements of verifying the application (inc appropriate grouping, classification etc) prior to commencing any work (Annex VII 4.2 & 4.3). 



    ------------------------------
    Erica Conway
    Vice President, IVD Regulatory Affairs - Europe
    Oxford
    United Kingdom
    ------------------------------



  • 4.  RE: EMDN Code (EU MDR Requirement)

    Posted 09-Sep-2022 08:53
    Edited by Kevin Randall 09-Sep-2022 08:55
    I agree that a Manufacturer's derivation of an EMDN code to help facilitate the NB's technical documentation assessment can be helpful.  In fact, when my firm prepares technical documentation files, an EMDN code(s) is one of the first things that appears in the technical documentation. That said, I think it's vitally important not to convolute a Notified Body's obligations with the Manufacturer's.  It's also important to know the difference between legislative mandates vs. guidance.

    Specifically for this discussion and Anon's question about the EU MDR:

    • A Manufacturer's legislatively-mandated use of the nomenclature codes is limited to the aforementioned Article 26 mandate tethered to the intent of the EUDAMED database.
    • While certain Manufacturers' technical documentation is subject to a conformity assessment of at least one representative device for each generic device group / category of devices [Article 52(4) & (6)], it remains true that this does not obligate the Manufacturer to assign an EMDN code for that purpose.
    • Instead, it is the NB (if any is involved) who is obligated to perform such sampling of the technical documentation.  For that, MDCG 2019-13 is a guidance intended for NBs to help NBs in deciding how to choose such sampling.  That guidance aims to leverage the EMDN coding to help the NB meet this obligation.  But that guidance is not a legislative mandate for the NB, and certainly not for a Manufacturer.
    • Guidelines like MDCG-13 are just that: guidelines.  They are not legislation, and we should all be careful not to ascribe legislative authority to such guidance.  For example, MDCG-13 (as with other MDCG guidelines) remind us of this in their disclaimer, "...The document is not a European Commission document and it cannot be regarded as reflecting the official position of the European Commission. Any views expressed in this document are not legally binding..."
    • I'm unaware of any particular obligations in EU MDR Annex VII 4.2 or 4.3 that mandate a Manufacturer to assign EMDN codes for the purpose of an NB's technical documentation assessment. Instead, these Annex VII provisions are standard obligations to be met by the NB and that govern the NB's control of the pre-conformity assessment application step. 


    ------------------------------
    Kevin Randall, ASQ CQA, RAC (Europe, U.S., Canada)
    Principal Consultant
    Ridgway, CO
    United States
    © Copyright 2022 by ComplianceAcuity, Inc. All rights reserved.
    ------------------------------



  • 5.  RE: EMDN Code (EU MDR Requirement)

    Posted 09-Sep-2022 09:13
    Edited by Erica Conway 09-Sep-2022 09:14
    Hi Kevin

    I agree that 'guidance is guidance', however, NBs are quite often been held to the requirements in those guidances by their regulators (designating CAs; and this can vary between Member State as to how the authority is viewing requirements). ... It is a long running discussion on the MDCG guidances being viewed like that.  I noticed this has come up as a point on application of guidances under MDCG 2022-14 recently (point 11) that 'notified bodies should be allowed more flexibility as to how to demonstrate compliance with legal requirements', admitting that it is an issue(?).
    This means, in the absence of any other clear practical way, the NBs will use the codes in their procedures to show how they are grouping devices (where applicable). So my point was that they may be practically needed at the time of application.

    ------------------------------
    Erica Conway
    Vice President, IVD Regulatory Affairs - Europe
    Oxford
    United Kingdom
    ------------------------------



  • 6.  RE: EMDN Code (EU MDR Requirement)

    Posted 10-Sep-2022 10:01
    Hello Anon

    Yes, you'll need the EMDN codes when you get around to EUDAMED data entry. You can see the list here:
    https://webgate.ec.europa.eu/dyna2/emdn/Z1206

    ------------------------------
    Anne LeBlanc
    United States
    ------------------------------