EU MDR Annex VI, Part C, Section 4.10 requires that the direct-marked UDI carrier for reusable devices which require cleaning, disinfection, sterilization or refurbishing between patient uses shall be permanent and readable throughout the intended lifetime of the device. Accordingly, if a direct-marked 2D DataMatrix UDI carrier fades to unreadability over the lifetime of the device, then such direct-marked UDI carrier would not meet the requirements of EU MDR Annex VI, Part C, Section 4.10.
Is the barcode to which you refer an additional UDI carrier providing redundancy with the 2D DataMatrix?
Is the barcode to which you refer direct-marked on the device?
------------------------------
Kevin Randall, ASQ CQA, RAC (Europe, U.S., Canada)
Principal Consultant
Ridgway, CO
United States
© Copyright 2023 by ComplianceAcuity, Inc. All rights reserved.
------------------------------
Original Message:
Sent: 07-Feb-2023 08:38
From: Anonymous Member
Subject: EU MDR 2017/745 UDI
This message was posted by a user wishing to remain anonymous
We as the manufacturer verify and grade the 2D direct part marked DataMatrix before we ship it out to the customer to ensure it meets the minimum requirements per GS1 Specification for UDI, a grade min 1.5. Over the lifetime of the device, the 2D DataMatrix may fade due to sterilization, cleaning, etc. The barcode is still scannable and can decode the 2D information for the life of the product . Does this meet the requirements of the EU MDR 745/2017 Part C The UDI System, Section 4.10?