  • 1.  EU MDR classification rule 21

    This message was posted by a user wishing to remain anonymous
    Posted 09-Nov-2022 10:42
    This message was posted by a user wishing to remain anonymous

    Hello RAPS community,

    Devices that are composed of substances or of combinations of substances that are intended to be introduced into the human body via a body orifice or applied to the skin and that are absorbed by or locally dispersed in the human body are classified as:

    class III if they, or their products of metabolism, are systemically absorbed by the human body in order to achieve the intended purpose;
    class III if they achieve their intended purpose in the stomach or lower gastrointestinal tract and they, or their products of metabolism, are systemically absorbed by the human body;
    class IIa if they are applied to the skin or if they are applied in the nasal or oral cavity as far as the pharynx, and achieve their intended purpose on those cavities; and
    class IIb in all other cases.


    Can you please help understand the above classification rule 21 from EU MDR? 
    1. Isn't body orifice part of skin? What's the difference here in mode/point of application? 
    2. Absorbed: does this apply to ANY kind of component? For ex, if a medical device is in form of a formulation that is composed majority of water, is water considered to be absorbed, and hence, the device in-turn is "absorbed"?
    3. Locally dispersed: Does this mean that the product is spilled around the surface of application, or it only means dispersed in terms of blood flow or via lymphatic system in the body?

    I appreciate your insight on the topic.

    Thank you,

    Another RAPS fellow



  • 2.  RE: EU MDR classification rule 21

    Posted 09-Nov-2022 11:21

    Dear Colleague,

     underneath please find my thoughts.

     

    1. Isn't body orifice part of skin? What's the difference here in mode/point of application?

    I think you may relay on Wikipedia Body orifice - Wikipedia. Human body orifices used for medical device applications are: mouth, ear canal, nostrils, vagina, anus. These places are mucosa, that eases the absorption of the medical devices ingredients, which means their entry into the lymphatic system and the blood stream. This means increased biological risk, to be covered in the risk management report.

    1. Absorbed: does this apply to ANY kind of component? For ex, if a medical device is in form of a formulation that is composed majority of water, is water considered to be absorbed, and hence, the device in-turn is "absorbed"?

    I think yes, any absorbed component may generate biological reaction (e.g. hypersensitivity, interactions with other chemicals, e.g active substances of medicines), this needs to be investigated in the clinical and risk management reports. There are medical devices for the reduction of the symptoms of sore throat or haemorrhoids. These may contain herb extracts or glycerine. These frequently listed as ingredients of medicinal products and get absorbed into the human body.

    1. Locally dispersed: Does this mean that the product is spilled around the surface of application, or it only means dispersed in terms of blood flow or via lymphatic system in the body?

    According to my understanding, locally dispersed means penetration and actual absorption through a barrier (skin or mucosa). You may find relevant details in the listed publications here dispersion of ingredients in the skin - Search Results - PubMed (nih.gov)  

    with regards



    ------------------------------
    Peter Mikó M.D
    ArtPharm Ltd.
    Gyermely
    Hungary
    ------------------------------



  • 3.  RE: EU MDR classification rule 21

    This message was posted by a user wishing to remain anonymous
    Posted 11-Nov-2022 08:09
    This message was posted by a user wishing to remain anonymous

    Thanks Peter! This is very elaborative and clear!


  • 4.  RE: EU MDR classification rule 21

    Posted 09-Nov-2022 12:03
    Edited by Kevin Randall 09-Nov-2022 12:45
    1. The EU MDR defines 'body orifice' to mean any natural opening in the body, as well as the external surface of the eyeball, or any permanent artificial opening, such as a stoma.

    2. In the context of medical devices, the Commission has traditionally interpreted wholly or mainly "absorbed" to mean the degradation of a material within the body and the metabolic elimination of the resulting degradation products from the body.  The MDCG (see MDCG 2021-24) has maintained this definition, except that they added "in the context of implantable devices".  Indeed, this definition of "absorbed" wasn't specifically intended for use in association with Rule 21, and I know of no further definition for that specific context.  However, the MDCG has stated for Rule 21 that manufacturers of substance-based devices should provide clear information supporting the mode of action through which the substance achieves the intended specific medical purpose as a basis for the application of this rule, including the site of application as and the site where the action is achieved in or on the body.  So you would need to see if the aforesaid definition for 'wholly or mainly absorbed', possibly with adaptations, would fit and could be leveraged for your device's mode of action.

    Regarding whether absorption of water would trigger, for example, the class III provisions of Rule 21, my personal opinion is that the presence of the device's other substances would essentially render the presence and absorption of the water a moot point.  The MDCG has said that this rule is for "exclusively substance-based" devices, and that 'substance' in this context means any "matter" that is part of the medical device.  From my degrees in chemistry and biology, water is a form of matter, and thus would, technically speaking, be included in the scope of this Rule.  But again, to me, that seems to be a moot point, as the device's other constituent substances are the primary concern for such a device.

    3. The MDCG has defined 'local dispersion' to mean the condition by which substances remain in a specific site without being distributed into the body via the blood and/or lymphatic system.

    Hope this helps.


    ------------------------------
    Kevin Randall, ASQ CQA, RAC (Europe, U.S., Canada)
    Principal Consultant
    Ridgway, CO
    United States
    © Copyright 2022 by ComplianceAcuity, Inc. All rights reserved.
    ------------------------------



  • 5.  RE: EU MDR classification rule 21

    This message was posted by a user wishing to remain anonymous
    Posted 11-Nov-2022 08:09
    This message was posted by a user wishing to remain anonymous

    Thank you Kevin! This is in addition to our current understanding.


  • 6.  RE: EU MDR classification rule 21

    Posted 09-Nov-2022 12:08

    The problem starts because the EU-MDR uses the technical term "substance" without definition. I generally use the definition in the REACH directive. (The medicinal product directive has a different definition, but it is more specific so not generally applicable.)

    Substance means a chemical element and its compounds in the natural state or obtained by any manufacturing process, including any additive necessary to preserve its stability and any impurity deriving from the process used, but excluding any solvent which may be separated without affecting the stability of the substance or changing its composition

    Mixture means a mixture or solution composed of two or more substances

    The EU-MDR Annex VIII defines a body orifice as any natural opening in the body, as well as the external surface of the eyeball, or any permanent artificial opening, such as a stoma. A body orifice is not skin, but an opening.



    ------------------------------
    Dan O'Leary CQA, CQE
    Swanzey NH
    United States
    ------------------------------



  • 7.  RE: EU MDR classification rule 21

    Posted 09-Nov-2022 12:49
    The MDCG has defined 'substance' to mean any matter that is part of the medical device, including those according to the definition of 'substance' in Article 1(3) of Directive 2001/83/EC, provided that they are not excluded by the MDR [e.g. exclusion criteria in article 1 (6) (h)].

    ------------------------------
    Kevin Randall, ASQ CQA, RAC (Europe, U.S., Canada)
    Principal Consultant
    Ridgway, CO
    United States
    © Copyright 2022 by ComplianceAcuity, Inc. All rights reserved.
    ------------------------------



  • 8.  RE: EU MDR classification rule 21

    This message was posted by a user wishing to remain anonymous
    Posted 11-Nov-2022 08:09
    This message was posted by a user wishing to remain anonymous

    Thank you for the feedback Dan!