Regulatory Open Forum

 View Only
  • 1.  EU MDR PMCF and Legacy Devices

    This message was posted by a user wishing to remain anonymous
    Posted 28-Feb-2023 08:38
    This message was posted by a user wishing to remain anonymous

    All,

    I have a "legacy device," class IIb, as defined in MDCG 2021-25. 

    The device has yet to be placed on market in the EU or even advertised, this is due to the business element of the company I am at. Additonally, I have procedures for all the transnational provisions in Article 120. Due to no devices being placed on market or advertised none of those processes have been utilized. Given this information would you expect to the company to have a PMCF plan drawn up or just procedures on how to create a PMCF plan? The same question can apply to the other processes in the Article 120 transitional provisions.

    Thank you for any insight provided.



  • 2.  RE: EU MDR PMCF and Legacy Devices

    Posted 28-Feb-2023 10:07
    Edited by Marcelo Antunes 28-Feb-2023 10:13

    You are required to implement and have all the procedures and records related to the regulatory requirements. 

    In a ideal world, as you are not marketing the device yet, it would be simple (and acceptable) to just justify why you do not have the records.

    However, in our world (when sometimes people just read too much into requirements and sometimes expect too much) it would be better to have the plan and related reports filled, even with no new information (you would still need to justify and make conclusions, anyway). 



    ------------------------------
    Marcelo Antunes
    Regulatory Strategy Consultant, HAS (Harmonised Standards) Consultant for the European Commission
    Jundiai
    Brazil
    ------------------------------



  • 3.  RE: EU MDR PMCF and Legacy Devices

    Posted 28-Feb-2023 15:30

    In order for the EU MDR Article 120 legacy device MDD certificate to remain valid, all of the applicable conditions upon which issuance of that MDD certificate was based shall be maintained regardless of whether devices have yet been placed on the market.  For example regarding the context of your particular question, Annex III requires a PMCF Plan (where not eligible for the EU MDR's general waiver opportunity) as a condition for maintenance of the validity of the legacy MDD certificate.  In other words, said PMCF Plan (and overarching PMS Plan) is generally required to be drawn up and ready for execution as a condition of maintaining the validity of the legacy MDD certificate.



    ------------------------------
    Kevin Randall, ASQ CQA, RAC (Europe, U.S., Canada)
    Principal Consultant
    Ridgway, CO
    United States
    © Copyright 2023 by ComplianceAcuity, Inc. All rights reserved.
    ------------------------------



  • 4.  RE: EU MDR PMCF and Legacy Devices

    Posted 28-Feb-2023 15:44
    Edited by Kevin Randall 28-Feb-2023 15:45

    Forgot to mention.  It is my understanding that, as a general rule, the planned/required PSUR shall also be drawn up at the required time.  If, at such time, no devices were yet placed on the market, then the PSUR simply states in the corresponding sections that there are yet no PMS / PMCF data collected for the marketing of the subject device because no subject devices have yet been placed on the market.

    However, be aware that part of the PMS plan upon which the legacy MDD certificate is contingent requires collection and analysis of PMS data not just on the marketing of the subject device.  Specifically, the PMS plan also requires collection and analysis of data from specialist or technical literature, databases and/or registers, and publicly available information about similar medical devices.  Be sure you have suitably addressed those required PMS aspects even if you haven't yet marketed the subject device.

    ------------------------------
    Kevin Randall, ASQ CQA, RAC (Europe, U.S., Canada)
    Principal Consultant
    Ridgway, CO
    United States
    © Copyright 2023 by ComplianceAcuity, Inc. All rights reserved.
    ------------------------------



  • 5.  RE: EU MDR PMCF and Legacy Devices

    Posted 01-Mar-2023 03:13

    Was going to comment don't forget about PSUR! but Kevin has gotten there!  Some good advice in the previous posts.



    ------------------------------
    Tom Patten,
    NSAI,
    Ireland.
    ------------------------------



  • 6.  RE: EU MDR PMCF and Legacy Devices

    This message was posted by a user wishing to remain anonymous
    Posted 01-Mar-2023 10:34
    This message was posted by a user wishing to remain anonymous

    All,

    Thank you for the input. Now hopefully I can get the owners to understand this in order to get resources to carry these items out.




  • 7.  RE: EU MDR PMCF and Legacy Devices

    Posted 02-Mar-2023 12:31
    Edited by Rajeswari Devanathan 02-Mar-2023 12:31

    Hi, 

    Article 120 of the MDR (Medical Device Regulation) outlines the provisions for the transitional period of the regulation's implementation. Even for device not launched but still it is certified, PMCF plan has to be drawn to make it complaint. 



    ------------------------------
    Raje Devanathan
    Amerisource Bergen
    TPIreg, Innomar Strategies
    Senior Manager - Regulatory Affairs, Medical Devices