Remember that "REF" per ISO 15223-1 / EN ISO 15223-1 (as amended) means catalog number. Accordingly, the UDI-PI shall not encode for the REF. Instead, the REF is encoded (if at all) within the UDI-DI. If the REF is encoded within the UDI-PI, then that would be a nonconforming UDI-PI leading to the need for significant and costly corrective actions, up to and including, device recall.
And if the label contains an expiry date or LOT, then the UDI-PI for either the U.S. or Europe shall (not should) encode for those production variables (notwithstanding U.S. class I device exceptions).
------------------------------
Kevin Randall, ASQ CQA, RAC (U.S., Europe, Canada)
Principal Consultant
Ridgway, CO
United States
© Copyright 2023 by ComplianceAcuity, Inc. All rights reserved.
------------------------------
Original Message:
Sent: 07-Aug-2023 02:16
From: Ronald Boumans
Subject: Expiration date AND/OR Manufacturing date.
In the MDR, Annex VI, Section 3.5 says: If a lot number, serial number, software identification or expiry date appears on the label, it shall be part of the UDI-PI. If there is also a manufacturing date on the label, it does not need to be included in the UDI-PI. If there is only a manufacturing date on the label, this shall be used as the UDI-PI.
For UDI, the answer to your question would be that it would be sufficient to add the expiry date to the label and include in the UDI-PI. If REF and LOT are also on the label, they should also be part of the UDI-PI.
------------------------------
Ronald Boumans
MDR Expert
Super PRRC
Netherlands
Original Message:
Sent: 03-Aug-2023 15:16
From: Anonymous Member
Subject: Expiration date AND/OR Manufacturing date.
This message was posted by a user wishing to remain anonymous
For a medical device with an expiration date, is it enough to indicate the expiration date in the product label, or should the manufacturing date be included as well?
REF , LOT, are included as well.
(for the US market)