Regulatory Open Forum

 View Only
  • 1.  FDA recall

    This message was posted by a user wishing to remain anonymous
    Posted 30-Dec-2022 08:37
    This message was posted by a user wishing to remain anonymous

    Hello All,

    I was wondering if anyone could help me guide with recall process for the FDA.

    1. Are there timelines as to in how many days manufacturers have to perform a recall? Are there due dates that needs to be met per the regulation for recall notification to both the FDA and  customer (consignee)?
    2. I see that per 21 CFR 806.10(d) does require manufacturers to submit to the FDA within 10-working days.

    Please provide any guidance on the requirements?

    Thank you!


  • 2.  RE: FDA recall

    Posted 30-Dec-2022 18:58
    Hello Anon,

    Yes, you are right. The report must be submitted to FDA within 10 working days from the time the firm initiates the correction or removal . If there is not a "risk to health" involved, a report to FDA is not required, but the manufacturer or importer must keep a record of the correction or removal.

    You can refer the below links for more details:

    Chapter 7 Recall Procedures (fda.gov)

    What is a Medical Device Recall? | FDA

    Recalls, Corrections and Removals (Devices) | FDA


    ------------------------------
    Ajit Basrur
    Worcester MA
    United States
    ------------------------------



  • 3.  RE: FDA recall

    Posted 30-Dec-2022 20:42
    Hi Anon

    The expectations may differ if you're recalling drugs, devices, foods, or whatever, but there's guidance for all on the website. Here are a couple more good starting points.

    https://www.fda.gov/drugs/cder-small-business-industry-assistance-sbia/key-elements-being-recall-ready

    https://www.fda.gov/safety/recalls-market-withdrawals-safety-alerts/industry-guidance-recalls

    ------------------------------
    Anne LeBlanc
    United States
    ------------------------------



  • 4.  RE: FDA recall

    Posted 31-Dec-2022 09:37

    It is important to distinguish between Corrections & Removals in Part 806 and recalls in Part 7.

    When your company decides to inform customer of the need for a correction or removal (C&R), then the Part 806 10-day clock starts to inform FDA. While the regulations don't specify a time to notify customers, don't delay. Your notifications should have gone out within the same 10-day window.

    Similarly, there is no timeline to receive notification that the customer has completed the action. If FDA classifies the C&R as a recall under Part 7 they will ask you for a recall plan with timing, depth of recall, etc.



    ------------------------------
    Dan O'Leary CQA, CQE
    Swanzey NH
    United States
    ------------------------------



  • 5.  RE: FDA recall

    This message was posted by a user wishing to remain anonymous
    Posted 03-Jan-2023 11:40
    This message was posted by a user wishing to remain anonymous

    Thank you all for the response!

    Hi Dan, 

    My understanding is that, if a manufacturer becomes aware of any issue, then they start the correction and removal as stated in Part 806. Notify FDA within 10-working days with the listed content on Section 806.10 (c) 1 – 13. FDA can always issue a recall based on the information submitted per 810?

    Also, if manufacturer decides that a recall is required. Then a voluntary recall is initiated per CFR 7. A recall strategy (sec 7.42) is drafted by the manufacturer and sent to FDA. A recall is initiated when the FDA is reviewing or even before. When drafting a recall strategy, it does mention that "result of health hazard evaluation" is to be taken into consideration, I thought that the HHE is performed by FDA and not by the manufacturer, is my assumption correct? If that is the case, what does the Manufacturer need to include in a recall strategy.

    I am also slightly confused as to what is the difference between "correction and removal" and "recall" aren't they considered the same?

    Thank you in advance!




  • 6.  RE: FDA recall

    Posted 03-Jan-2023 16:55

    There is a particular process to follow. First, Part 806 is for devices only and the manufacturer implements it. Part 7 is for all FDA regulated products and FDA implements it. Part 806 may feed information to Part 7.

    Start by deciding whether your company has a device issue about which you should inform your customers. Determine what the customer' actions and inform them. Send the information to, at least, the initial consignees using the data from 820.160(b) without "undue haste". (In ISO 13485:2016 this is an advisory notice.)

    Determine if the issue is exempt from reporting by using 806.1(b). In addition, there is another exemption in the post-market cybersecurity document.

    If it is exempt from reporting, produce the 806.20 documentation as quality records. You do not have to report to FDA that the issue is exempt from reporting. STOP.

    If it is not exempt from reporting determine if it is a correction, a removal, or both using the definitions in 806(2). An issue could both depending on who implements the action. A Field Service Rep may handle some customers on site (correction) while other customers need to send the device to another location for implementation (removal). This distinction determines the format of the report numbers.

    Submit the reports using eSubmitter.

    FDA will determine whether the report is a recall and its class. They will also tell you things like the depth of the recall and the effectiveness. You will prepare a recall plan and submit it to FDA.

    The important thing is the manufacturer does not decide if the action is a recall. FDA makes that decision based on the information submitted under 806.10.

    FDA conducts the HHE to help make the decision on whether the action is a recall and its class. The HHE helps guide the manufacturer's recall strategy for things like the depth of the recall and the effectiveness check.

    Many people use the term "correction and removal" as well the term "recall" as synonyms.

    Compare the definition of correction in Part 7 with the definition in Part 806.

    Compare the definition of removal in Part 7 with the definition in Part 806.

    Read the definition of recall in Part 7 and notice that the term is not in Part 806.



    ------------------------------
    Dan O'Leary CQA, CQE
    Swanzey NH
    United States
    ------------------------------



  • 7.  RE: FDA recall

    This message was posted by a user wishing to remain anonymous
    Posted 04-Jan-2023 14:01
    This message was posted by a user wishing to remain anonymous

    Thank you Dan for the detail explanation and your time!

    What about 21 CFR 7.46 which is Firm-initiated recall. When does this apply to the manufacturer? I believe this is applicable when a manufacturer voluntarily performs a recall and notifies to the FDA with the set of information addressed in the regulations.

    Thank you!


  • 8.  RE: FDA recall

    Posted 04-Jan-2023 17:50

    7.46 has a very specific set of circumstances:
    The firm decides to perform a correction or removal

    If the firm believes the product is violative, then immediately notify FDA

    It is a recall only if FDA agrees that it is a violation subject to legal action

     

    When you read through the rest of the section, it sounds a lot like Part 806 for devices. This is no coincidence.

    Remember that Part 7 covers all FDA regulated products such as mayonnaise, shell eggs, tanning booths etc. Part 806 is  specific to medical devices. In practice when a device manufacturer initiates a C&R under Part 806 FDA considers it under 7.46. If you look at some of the device recall reports, you will often see the phrase "voluntary recall".



    ------------------------------
    Dan O'Leary CQA, CQE
    Swanzey NH
    United States
    ------------------------------