Hello,
Typically if you are making a finished medical device the answer would be no. Good Laboratory Practices (GLP) are usually required for compliance of organisations doing testing of components, materials, or other non-patient related results, i.e. biocompatibility, sterilisation, shelf-life, animal, etc. Depending on the IVD, there may be aspects of GLP to consider in the manufacturing and/or testing of the medical device, but this also is commonly applicable during design and development such as verification and validation. This might not necessarily be applicable during routine manufacturing, though again unless there are involved laboratory tests being done in-house. However, generally the manufacturing and testing of a finished medical device would follow GMP requirements.
------------------------------
Richard Vincins ASQ-CQA, MTOPRA, RAC
Vice President Global Regulatory Affairs
------------------------------
Original Message:
Sent: 09-Jan-2023 19:52
From: Anonymous Member
Subject: GLP Regulations
This message was posted by a user wishing to remain anonymous
Hello everyone,
I have a question that I would appreciate your advice on.
we are an IVD medical devices company manufacturing lateral flow tests. do we need to comply with GLP regulation?
kind Regards, Asma