Hi Kevin
So far it seems FDA still considers all point-of-care testing to be high enough risk to warrant premarket notification/approval. Maybe it's less likely that a test done in a lab would return wrong results leading to improper treatment? Even if the test itself seems innocuous, there doesn't seem to be any exception, apart from emergency use authorizations.
https://www.fda.gov/files/medical%20devices/published/Procedures-for-Class-II-Device-Exemptions-from-Premarket-Notification--Guidance-for-Industry-and-CDRH-Staff-%28PDF-Version%29.pdf
https://www.fda.gov/medical-devices/ivd-regulatory-assistance/clia-categorizations
https://ascls.org/point-of-care/
https://www.fda.gov/regulatory-information/search-fda-guidance-documents/transition-plan-medical-devices-issued-emergency-use-authorizations-euas-related-coronavirus-disease
------------------------------
Anne LeBlanc
United States
------------------------------
Original Message:
Sent: 21-Dec-2023 09:25
From: Kevin McLellan
Subject: Interpretation of Section 862.9, Limitation to 510k Exemptions
Hi....I'd like to confirm my interpretation of the FDA regulation in section 862.9, the Limitations of the 510k Exemption. Anyone care to weigh in? It's at 21 CFR Part 862, Subpart A, Sec 852.9.
It begins:
"The exemption from the requirement of premarket notification . . . for a generic type of class I or II device is only to the extent that the device has existing or reasonably foreseeable characteristics of commercially distributed devices within that generic type or, in the case of in vitro diagnostic devices, only to the extent that misdiagnosis as a result of using the device would not be associated with high morbidity or mortality". (emphasis me)
I'm reading this to mean that the exceptions itemized below this (items A-C(1-9)), only apply when the device (in my case an in vitro diagnostic device), has risks "associated w high morbidity or mortality,". In other words, the exceptions don't stand on their own, but apply only under this high-risk scenario.
At the end of the exceptions list, items C(8-9) appear to include all non-invasive and all near-patient in vitro testing, but it doesn't make sense to exclude ALL these, but only in the high-risk scenarios. That's how I read this.
Any feedback appreciated!
------------------------------
Kevin McLellan
Owner
Memphis TN
United States
------------------------------