This is an area that can be confusing.
First, it is not clear if your company's role is to run the test on patient samples (a lab) or to manufacture the test (IVD manufacturer) for use in a lab. Also, is it a lab in your company, a lab in another company, or both.
I'll infer your company is the manufacturer. First, start with 812.2(c)(3) to determine if your IVD is exempt from the IDE regulations. This will take you to 809.10(c).
If the IVD is not exempt, then you will need to file an IDE following 812.
If the IVD is exempt, determine if your project is in either the RUO phase or the IUO phase of development.
In either case, determine whether the IVD is Class I. If so, design controls under 820 do apply.
Based on the description, your IVD cannot be an LDT, because it is transferred from another site. It won't pass the "four walls" test, which means an LDT must be developed and run in the same lab. Transfer is not allowed. Consequently, the CLIA requirements for developing an LDT in a high-complexity lab do not apply.
As an aside, FDA will announce its plans to regulate LDTs in August this year.
------------------------------
Dan O'Leary CQA, CQE
Swanzey NH
United States
------------------------------
Original Message:
Sent: 20-Jun-2023 17:35
From: Joscelyn Bowersock
Subject: Laboratory Developed Tests for IDE Submission
Hello all,
I am working with an European company that is transferring a test to my site for FiH testing. My regulatory group is looking at 21CFR820 for design control and doesn't quite understand CLIA or laboratory developed testing design controls. Does anyone here have experience with submitting an LDT for an IDE? I have some experience (developing treatment-related testing), but am not sure if anyone has good wordsmithing for describing CLIA LDT requirements in EU speak.
------------------------------
Joscelyn Bowersock MSc
Quality and Regulatory
Mint Hill NC
United States
joscelyneb@outlook.com
------------------------------