Hello Anon,
The structure and content of the Manufacturing section of Technical Documentation for EU MDR compliance can vary significantly from device type to device type - and by the classification of the device. It is more important to think of the Technical Documentation as an index or a structured approach for presentation of detailed information. For the manufacturing process description this could be a narrative description, a tabular listing of steps in the process, or a flow chart with some explanation. It is fairly high level in the Technical Documentation index file - again what is important is then referencing documents such as the Device Master Record/Medical Device File, work instructions, testing procedures, etc., which can then review further detail.
The Bill of Material (BOM) in the Technical Documentation index file is either a reference to the listing which is in further detail or providing the first or second level of the BOM. Putting in the whole BOM in the Technical Documentation index/file might be too much unless it is a less complex device. It really depends though, because some electronic devices may need list of crucial components in the Technical Documentation or the composition of a medical device provided - there is no specific rule or guideline for what is in there. Read the Technical Documentation file from the reviewers perspective to make sure it is understandable, clear, and has references to the further detail. Then when submitting the Technical Documentation, work with your Notified Body to understand what level of documentation needs to be provided initially - such as providing the full indented BOM as a separate document in the submission.
------------------------------
Richard Vincins ASQ-CQA, MTOPRA, RAC
Vice President Global Regulatory Affairs
------------------------------
Original Message:
Sent: 01-Aug-2023 01:46
From: Ronald Boumans
Subject: MDR - Device BOM, Production
In addition to the reply by Rajeswari, I would also suggest using a risk based approach. There are three main risks that should be considered:
- Biocompatibility; parts that come into contact with users/patients must be fully identified and they must be addressed in the biocompatibility evaluation.
- Device failure; if the notified body wants to verify the parts in the risk analysis that cover device failure, they need to understand how it is build.
- Hazardous substances and waste management; the presence of certain substances may trigger other legislation and the notified body needs to be able to evaluate that. Therefore you must either go down to the smallest parts of the device, or have components or sub assemblies covered by relevant certificates.
------------------------------
Ronald Boumans
MDR Expert
Super PRRC
Netherlands
Original Message:
Sent: 30-Jul-2023 18:48
From: Anonymous Member
Subject: MDR - Device BOM, Production
This message was posted by a user wishing to remain anonymous
Hi forum,
Please advise.
We are in the process of compiling the technical documentation, and needed clarity around Device BOM- what is the expectation here? Is it only to include the BOMS at the top level - assemblies and subassemblies or is it required at the component level, we have an active device.
For the manufacturing section, if there are no critical processes involved, how detailed should the manufacturing process flow chart and procedures be?
thanks