Q1: As I understand it, if the Article 22 Procedure Pack Producer (PPP) doesn't carry out the sterilization in accordance with the Procedure Pack (PP) constituent device manufacturer's instructions, then the PPP is not allowed to apply Article 22(3). A key reason for this is that, without consulting the manufacturer(s) of the PP constituent device(s) and thereby not following the constituent device manufacturer's sterilization instructions, the constituent device integrity or other constituent device GSPR conformity could be adversely affected. This is because the Article 22(3) route doesn't involve assuring full GSPR conformity of the PP constituents. Instead, the Article 22(3) route and the PPP's corresponding Annex IX or XI conformity assessments made thereunder are for the procedure pack as a whole and limited only to assuring sterility. Thus, while a PPP's Article 22(3) route may indeed assure sterility, it doesn't involve assuring the maintenance of various other GSPR requirements (e.g., constituent device stability/reliability, constituent device performance, constituent device labeling integrity/readability, etc.) for the resterilized PP constituent devices.
Q2) The reason the Article 22(3) route allows an abbreviated conformity assessment (focusing only on sterility) is because it is presumed that the PP constituent device manufacturers have assured that the full GSPR conformity of their constituent device(s) will be maintained upon repeated sterilization. If the constituent device manufacturers don't provide such assurance, then it leaves a profound gap in GSPR conformity assurance of the PP's constituent devices. Thus, either the constituent device manufacturers must fill that gap in support of the PPP's Article 22(3) route, or else the PPP is required to instead apply Article 22(4) so as to assure full GSPR conformity of the Procedure Pack and all of its constituents by applying a full Article 52 conformity assessment that is not limited to just sterility.
------------------------------
Kevin Randall, ASQ CQA, RAC (Europe, U.S., Canada)
Principal Consultant
Ridgway, CO
United States
© Copyright by ComplianceAcuity, Inc. All rights reserved.
------------------------------
Original Message:
Sent: 06-Feb-2024 05:00
From: Anonymous Member
Subject: Procedure Pack Producer & Article 22 MDR.
This message was posted by a user wishing to remain anonymous
Dear community,
this is an open question regarding the interpretation of Article 22 of the MDR and more in particular the 22(3). See below a case study and thx in advance for you comments and thoughts.
I'm the Legal Manufacturer of a Class III, Sterile Medical Device (MD-A). For the conformity assessment, i've validated my device for 1-year Shelf-Life after 1x Cycle of Sterilization (Gamma irradiation). My device is CE-mark per EU MDR. I've been approached by a Procedure Pack Producer (PPP), because they are willing to place my device (MD-A) as part of a Procedure Pack, with other medical devices. All other medical devices are also CE-marked.
The PPP is willing to Sterilize the Procedure Pack (meaning double sterilization for my device MD-A). Per MDR 22(3) the PPP shall draw-up a statement declaring that sterilsation has been carried out in accordance with the manufacturer's instructions.
Questions:
Q1) If, as the Legal manufacturer of MD-A, I'm NOT providing the sterilisation details/ instructions/ specifications of my medical device to the PPP, can the PPP apply Article 22(3)? or the Procedure Pack manufacturer should follow Art 52, as my device will be "double sterilized" following their validated sterilization process and parameters?
Q2) If i'm willing to provide the sterilsation instructions/ specs/ of my device (MD-A) to the PPP and the PPP is willing to sterilise the whole procedure pack (total 2 sterilization cycles), Am I (as the legal manufacturer of MD-A) responsible to "validate the dual sterilization" and shelf life before placing the device as part of the Procedure Pack?
Thanks in advance for your thoughts.
BR