  • 1.  REACH application to medical device

    This message was posted by a user wishing to remain anonymous
    Posted 01-Dec-2023 09:28
    This message was posted by a user wishing to remain anonymous

    Dear colleagues,

    We have been informed by the supplier of the main component of one of our devices, that the material contains a SVHC (cat. 1B) in a concentration lower than the maximum limit allowed by the REACH regulation (0.3%).

    Based on our calculations, the maximum amount of this SVHC in our device would be 0.2% w/w.

    Since our device is a mixture of substances, according to REACH, is the maximum limit of 0.3% applicable also to the finished device?

    Reg (EU) 2017/745 when there are substances of category 1A or 1B in a concentration higher than 0.1%, we have to provide additional information (e.g. rationale on the potential patient exposure etc) and the related labelling: is this applicable in our case? 

    Since there are discrepancies between regulations, we are discussing with other departments which one has the priority.

    Thanks in advance.



  • 2.  RE: REACH application to medical device

    Posted 01-Dec-2023 13:34

    Dear Anon

    The short answer is yes. With more than 0.1% (w/w), there may be communication requirements. I'm curious about the discrepancies you mentioned.

    The guidance, or the guidance-in-a-nutshell, on substances in articles, may be helpful...

    https://echa.europa.eu/documents/10162/2324906/articles_en.pdf

    https://echa.europa.eu/documents/10162/2324906/nutshell_guidance_articles2_en.pdf



    ------------------------------
    Anne LeBlanc
    United States
    ------------------------------



  • 3.  RE: REACH application to medical device

    Posted 02-Dec-2023 12:35

    The MDR and REACH are separately applicable and do not have any overlap or borderline provisions vis-a-vis each other.



    ------------------------------
    Erik Vollebregt
    Partner
    Amsterdam
    Netherlands
    ------------------------------



  • 4.  RE: REACH application to medical device

    Posted 03-Dec-2023 18:36

    I agree with Anne and Eric comments.  In general, the EU has many additional directives that needs to be applied for medical devices depend on the types.

    Such as 

    RoHS

    Machinery directive

    WEEE



    ------------------------------
    Armin Beck
    CEO
    SunTrixConsulting LLC
    El Dorado Hills, CA