Shahram,
There are many factors which go into inspection and testing of in-process product and finished product when using a sampling approach. There are sampling methods and plans which can be applied such as ISO 2859 or the newer ISO 2859x series or the older military standard which is still used today. However, sampling of finished goods does depend on many factors such as if the test and/or inspection is visual, qualitative, quantitative, using automated system, manual, etc. When determining if sampling is needed or can be done this also should rely on the verification and validation testing which has been performed of the device to determine if there are things like CTQ - Critical to Quality attributes. You might want to seek some expert advice or research into this as there are many articles and books on sampling methods, but it would be specific to the device.
------------------------------
Richard Vincins ASQ-CQA, MTOPRA, RAC
Vice President Global Regulatory Affairs
------------------------------
Original Message:
Sent: 14-Apr-2023 18:57
From: Dan O'Leary
Subject: Sampling method in finish product inspection
Using a sampling plan at final inspection is acceptable with a caveat. Device class is not a factor.
If you don't fully verify the process output, such as by using a sampling plan instead of 100% inspection, then you must validate the process.
There are two views of 100% inspection. In one view, the predominant view, 100% inspection means checking every product for the critical to quality, CQT, characteristics. I've had people say that 100% inspection requires checking every product for every characteristic.
You didn't say anything about the sampling plan, so I'll infer it is a lot acceptance plan. In the US, the common plans are Z1.4 and c=0. The c=0 plans have a strange OC curve such that the probability of acceptance drops off quickly. I've seen cases where poor quality requires rejection of almost every lot. In this case you are back to 100% inspection to screen the lot.
Both Z1.4 and c=0 have provisions for switching rules. I recommend you implement them since they can dynamically adjust the sampling plan. A good quality history can reduce the sample size and still provide the desired level of protection.
------------------------------
Dan O'Leary CQA, CQE
Swanzey NH
United States
Original Message:
Sent: 14-Apr-2023 16:40
From: Shahram Ahmadi
Subject: Sampling method in finish product inspection
Hi,
I have a question regarding the sampling method in the final inspection step of class II MDs (product realization) in manufacturing premises. Is it acceptable? For example, AQL=0.25
Kr,