Hi Nawal,
There is a process to follow, which I'll briefly outline. The process is in two regulatory requirements: FDA QSR and ISO 13485:2016. QSR is clearer, but in the next few years, FDA will adopt the 13485 version.
The process steps to define the requirements for the supplier, evaluate potential suppliers, select the supplier that is best fit, periodically reevaluate all suppliers against a set of criteria.
Define the requirements
List all the things that are important for this supplier. Examples include capability, capacity, and landed cost. Look at regulatory issues such as FDA 483s, OSHA citations, EPA citations, etc. Consider whether the supplier needs a QMS such as ISO 9001, FDA QSR, or ISO 13485.
Identify three (any number will work, but three is tradition) potential suppliers and ask each about the criteria defined. This could be a simple set of questions. If written in a form, it should be two pages or less.
Score each supplier's response to each item on a scale of 1 to 5. 3 means the supplier meets the requirement; 4 or 5 exceeds the requirement. Plot the results on a radar chart using MS Excel. The higher scoring suppliers will be further out on the chart.
Select the supplier that is the best fit for your company and document the reason. Add that supplier to the Approved Supplier List.
Periodically re-evaluate each supplier. I recommend using four criteria: on time delivery, lot acceptance rate at incoming, adequate QMS, and ease of working with the supplier. Each supplier's evaluation should take no more than five minutes. If a supplier doesn't meet all the criteria, then do a separate analysis.
Keep records of everything you do because the regulators will want to see them.
For more detail, find the GHTF guidance documents, because they are great. Because GHTF doesn't exist anymore, they moved to the IMDRF website.
Regards,
Dan
------------------------------
Dan O'Leary CQA, CQE
Swanzey NH
United States
------------------------------
Original Message:
Sent: 19-Apr-2023 15:38
From: Nawal Tiouri
Subject: Supplier Evaluation
Hi Dan ,
The company where I work is still in the development stage of a medical device, so for now we are using a random of suppliers online. However, in the near future we must start selecting and evaluating our suppliers, so my question is: Is there a process to follow? Thank you so much for your response.
Best,
Nawal
------------------------------
Nawal Tiouri
Ypsilanti MI
United States
Original Message:
Sent: 19-Apr-2023 13:48
From: Dan O'Leary
Subject: Supplier Evaluation
My recommendation is that you don't categorize your suppliers this way. There is no regulatory requirement that you do, so it becomes a non-value added to classify and maintain the classification. Having done a lot of supplier management consulting work, I've seen the classification system then result in activities with frequency depending on the classification.
One of the worst cases is frequent audits of critical suppliers. Usually, the manufacturer doesn't have the resources to conduct the supplier audits, they are missed, and in the worst case results in a Warning Letter for failure to follow your own procedures. Instead of supplier audits, ask your suppliers for any third-party audit reports. You have probably made it part of the T&Cs, so should be willing to give you the reports. Explain that you are in a regulated industry and must evaluate and re-evaluate suppliers. Give the supplier two choices – provide the audit report under confidentiality or your company will have to duplicate the audit cost the supplier additional time, resources, etc. Usually, the supplier will take the first choice.
As part of reevaluation collect metrics on on-time delivery and lot acceptance rate at incoming. Set standards for acceptable performance.
There is one caveat to classification. If your device has the MDD or MDR CE Mark issued by an NB, then your suppliers may be subject to NB unannounced audits. Often the NB will ask for your critical suppliers. Check the definitions in your contract or other NB documentation. This is a classification for the NB's use, not yours.
If you still feel compelled to classify suppliers this way, use the definitions from the GHTF series on supplier management.<o:p></o:p>
------------------------------
Dan O'Leary CQA, CQE
Swanzey NH
United States
Original Message:
Sent: 19-Apr-2023 13:24
From: Nawal Tiouri
Subject: Supplier Evaluation
Hello everyone. I'm thrilled that I joined RAPS, I have a question about the supplier assessment. We are implementing a new SOP for the supplier evaluation, so I'm wondering how to categorize our suppliers from critical to non-critical. Please share your thoughts and feedback with me.
------------------------------
Nawal Tiouri
Ypsilanti MI
United States
------------------------------