Original Message:
Sent: 23-Feb-2024 04:49
From: Edward Ball
Subject: Universal Precautions
Dan, not sure I can add much more to Richard's great reply. My instinct was to go to the MHRA's Managing Medical Devices guidance, as it used to contain information on reporting adverse incidents and how to handle devices under investigation (or so I recalled, it has been a long time since I referred to it regularly)...but it does not. It does have some guidance on decontamination but it is more general about when in-use etc. It does however point to these NHS guidance documents for decontamination in general:
You probably already have this covered, but it is worth remembering to ensure that sufficient evidence of the device's state before cleaning and then decontamination is gathered (e.g. photos, videos). Back in my MHRA days, we would obtain as much as we could safely before we then did any cleaning and decontamination. And the method of cleaning, decontamination should probably recorded so that there is evidence of what was done to the device. This depends on the incident and the potential for any criminal or legal proceedings.
------------------------------
Ed Ball
Manager, Intelligence & Innovation
United Kingdom
Original Message:
Sent: 22-Feb-2024 07:31
From: Dan O'Leary
Subject: Universal Precautions
Richard,
Thank you. This is very helpful.
Dan
------------------------------
Dan O'Leary CQA, CQE
Swanzey NH
United States
Original Message:
Sent: 21-Feb-2024 03:01
From: Richard Vincins
Subject: Universal Precautions
Hello Dan,
Personally, I still use the combination of the Bloodborne Pathogen standard and OSHA's requirements under 29 CFR 1910 for handling of returned goods with potential exposure: https://www.cdc.gov/niosh/learning/safetyculturehc/module-5/5.html#:~:text=The%20Bloodborne%20Pathogen%20Standard%3A&text=Requires%20the%20use%20of%20Universal,infectious%2C%20to%20protect%20against%20pathogens. and https://www.osha.gov/bloodborne-pathogens/worker-protections which are considered the most stringent rules in US for worker exposure. There are some guidance papers and procedures written around these which are not necessarily on the OSHA or CDC website. As an example, I know the cleaning and disinfecting process is a little different between hospitals in different States and also how companies manage exposure. I think the only comprehensive guides you will find are published by other association groups, private entities, or individual hospitals (https://www.dhs.wisconsin.gov/hai/precautions.htm as an example). The procedures I write now for handling returned goods was based off one of the university hospitals procedure - I do not remember which one now - but it was fairly comprehensive and written well.
------------------------------
Richard Vincins ASQ-CQA, MTOPRA, RAC
Principal Strategy Consultant
NAMSA
Original Message:
Sent: 20-Feb-2024 19:54
From: Dan O'Leary
Subject: Universal Precautions
I'm working on a project for a client, a US medical device manufacturer, who could have received returned product that has had patient contact.
I have not worked in this area for a few years, so I'm not sure I'm up to date. Then, the situation was covered under universal precautions which typically included:
Decontamination of returned product at the returning site before the return
Personal protective equipment, PPE, for any person handling returned product
Segregation of returned product to prevent contamination of new product
Decontamination at the receiving site, the device manufacturer, before investigation
As I recall, for example, FDA recommended certain disinfectants recommended by OSHA. Many of the requirements focused on blood-borne pathogens.
In my client's case it is unlikely the returned product has blood contamination, but universal precautions assume this kind of contamination.
What is the current state of recommendations? Is there a comprehensive guide for US based medical device manufacturers?
------------------------------
Dan O'Leary CQA, CQE
Swanzey NH
United States
------------------------------