Dear Anon:
Not knowing what device type you are inquiring about-here below are more FDA Guidance that may possibly help you with repeat testing from human factors use failures including design or labeling.
See attachment here for additional LIST of FDA PDF Guidance discovered with SmartSearch beyond your post-referenced guidance below:
"Applying Human Factors and Usability Engineering to Medical Devices"
SmartSearch-our Free, no Credit Card gimmick is a full-text, word by word FDA content (CFRs and US FDA Guidance) search hosted at our secure Microsoft authenticated FDA content portal site. We searched full-text (human* NEAR factor* ) NEAR (design* AND label*) ALL 638 FDA PDF Guidance hosted on FDA.GOV website.
Request for Free limited access 30days-no card gimmicks when requesting to RAPs Members here:
https://estarhelper.com/request-free-smartsearch-estarhelper/#estar π
SmartSearch collects FDA content (public domain-no copyrights) and render them full-text search enabled UNLIKE FDA.gov website's meta-data based search page (* if not in file name or title or description-guidance list is incomplete and PDF missed from FDA.GOV search results when matching your keywords*).
See latest Linkedin post for more - https://tinyurl.com/krt2xm5f π
Best regards.
Ram Balani eSTARHelper LLC
------------------------------
Ram Balani
CEO
FDASmart Inc. /eSTARHelper LLC
www.estarhelper.comAmawalk , New York
rbalani@fdasmart.com2019130558
https://tinyurl.com/2wkxp69yon US FDA eSTAR for 510(K)
------------------------------
Original Message:
Sent: 29-Jul-2022 06:35
From: Jonathan Amaya-Hodges
Subject: Usability Summative Testing / Re-Testing
Anonymous,
Generally, yes, this requires 're-validation' (i.e. another Human Factors study) to address the changes made in response to the original results. However, it does not have to be a wholesale repeat of the initial study - instead, it could focus on only the relevant portions affected (or potentially affected) by the changes. When coupled with the original results, it could provide support for design validation of the overall product (assuming generally positive data). This could help reduce the duration/cost of the second study (though possibly not size, unless only specific user group(s) were affected in which case those could be reduced).
Feel free to reach out directly if you require more specific advice.
------------------------------
Jonathan Amaya-Hodges
Director, Technical Services
Sharon MA
United States
Original Message:
Sent: 28-Jul-2022 04:26
From: Anonymous Member
Subject: Usability Summative Testing / Re-Testing
This message was posted by a user wishing to remain anonymous
Dear Community,
"If the results of human factors validation testing include use errors on critical tasks or participant feedback indicating difficulty with critical tasks, stating in a premarket submission that you mitigated the risks by modifying the instructions for use or some other element of labeling is not acceptable unless you provide additional test data demonstrating that the modified elements were effective in reducing the risks to acceptable levels." - FDA Guidance (Applying Human Factors and Usability Engineering to Medical Devices)
From your experience, could you please help me understand whether this implies that the summative testing must be repeated with recruited participants (quite costly)? Is there any other acceptable way of validating the mitigations?
I appreciate your time and help.