    Dear Community,

    "If the results of human factors validation testing include use errors on critical tasks or participant feedback indicating difficulty with critical tasks, stating in a premarket submission that you mitigated the risks by modifying the instructions for use or some other element of labeling is not acceptable unless you provide additional test data demonstrating that the modified elements were effective in reducing the risks to acceptable levels." - FDA Guidance (Applying Human Factors and Usability Engineering to Medical Devices)

    From your experience, could you please help me understand whether this implies that the summative testing must be repeated with recruited participants (quite costly)? Is there any other acceptable way of validating the mitigations?
     
    I appreciate your time and help.

    Best regards


  • 2.  RE: Usability Summative Testing / Re-Testing

    Posted 29-Jul-2022 06:35

    Anonymous,

    Generally, yes, this requires 're-validation' (i.e. another Human Factors study) to address the changes made in response to the original results. However, it does not have to be a wholesale repeat of the initial study - instead, it could focus on only the relevant portions affected (or potentially affected) by the changes. When coupled with the original results, it could provide support for design validation of the overall product (assuming generally positive data). This could help reduce the duration/cost of the second study (though possibly not size, unless only specific user group(s) were affected in which case those could be reduced).

    Feel free to reach out directly if you require more specific advice.



    ------------------------------
    Jonathan Amaya-Hodges
    Director, Technical Services
    Sharon MA
    United States
    ------------------------------



  • 3.  RE: Usability Summative Testing / Re-Testing

    Posted 29-Jul-2022 09:05
      |   view attached

    Dear Anon:


    Not knowing what device type you are inquiring about-here below are more FDA Guidance that may possibly help you with repeat testing from human factors use failures including design or labeling.

    See attachment here for additional LIST of FDA PDF Guidance discovered with SmartSearch beyond your post-referenced guidance below:

     "Applying Human Factors and Usability Engineering to Medical Devices"

     SmartSearch-our Free, no Credit Card gimmick is a full-text, word by word FDA content (CFRs and US FDA Guidance) search hosted at our secure Microsoft authenticated FDA content portal site. We searched full-text (human* NEAR factor* ) NEAR (design* AND label*) ALL 638 FDA PDF Guidance hosted on FDA.GOV website.

    Request for Free limited access 30days-no card gimmicks when requesting to RAPs Members here:

    https://estarhelper.com/request-free-smartsearch-estarhelper/#estar  πŸ‘€


    SmartSearch collects FDA content (public domain-no copyrights) and render them full-text search enabled UNLIKE FDA.gov website's meta-data based search  page (* if not in file name or title or description-guidance list is incomplete and PDF missed from FDA.GOV search results when matching your keywords*).

    See latest Linkedin post for more - https://tinyurl.com/krt2xm5f πŸ“š

     Best regards.
    Ram Balani eSTARHelper LLC



    ------------------------------
    Ram Balani
    CEO
    FDASmart Inc. /eSTARHelper LLC www.estarhelper.com
    Amawalk , New York
    rbalani@fdasmart.com
    2019130558
    https://tinyurl.com/2wkxp69y
    on US FDA eSTAR for 510(K)
    ------------------------------



  • 4.  RE: Usability Summative Testing / Re-Testing

    Posted 29-Jul-2022 08:01

    The first issue is the wording of the FDA guidance document. The manufacturer does not "mitigate" risks but reduces them. Reduce has the sense of preventing the harm with mitigate has the sense of working on the effect of a harm on the patient or user. ISO 14971:2019 does not contain the word mitigate.

    The paragraph deals with critical tasks. If you find a problem, then you need to fix it. However, there is still the issue of knowing the fix works. This means retesting.

    While the paragraph says to provide additional test data, the presumption is to run part of the summative evaluation. It does not need to be the full set and could have a different sample size. For example, if one category of users experience the problem, then you probably don't need representatives from the other categories. Write a simple test plan for the retest.

    Lastly, FDA does not care about your cost to acquire data, only that you provide good data.



    ------------------------------
    Dan O'Leary CQA, CQE
    Swanzey NH