    Is it always considered subvisible particles in injectable vials particles that are smaller than 100um?
    What would be the USP and EP limits for subvisible particles?

    Kind regards



  • 2.  RE: Visible particles

    Posted 10-May-2023 17:40

    100 µm - 150 µm size is usually considered to be the size limit between sub-visible and visible particles.
    Note that is it valid for isotropic particles - for longitudinal particles (fiber-like), the size limit can go up to 300 µm.
    These considerations can be found in USP <1790>.
    And in other documents if it is associated to Visual Inspection of other components/assemblies than the classical glass vials / syringes.