If the new Class I device is, by FDA regulation, GMP-Exempt, then 21 CFR 820.180 and 198 as applicable are the only mandatory sections of Part 820.
If the new Class I device is
not GMP-Exempt, then all applicable sections of Part 820 are mandatory except for 820.30 unless the Class I device is one of those encompassed by 820.30(a)(2).
FDA's Part 820 requires risk "analysis" for those devices that aren't exempt from 820.30 design controls. FDA's latest preferred approach for risk "analysis" is indeed a more holistic ISO 14971 risk "management" approach. Yet ISO 14971 is not mandatory as a means for risk analysis. In other words, if it's appropriate for a Class I (low-risk) device to only apply FMEA rather than a broader ISO 14971 technique like PHA, then that is not prohibited by the FDA for such Class I device.
Also, be careful about declarations that risk management and/or analysis aren't required for Class I GMP-Exempt or Design-Control-Exempt devices. For example, it isn't really possible to execute proper complaint handling (820.198) without a device risk management scheme in place. Likewise, it isn't really possible to properly handle deliberations about recalls and Part 806 reporting without a device risk management scheme in place. Moreover, when promulgating Part 820, FDA repeatedly made clear that the extent of documentation and solutions employed for meeting applicable Part 820 elements is to be based on risk.
------------------------------
Kevin Randall, ASQ CQA, RAC (U.S., Europe, Canada)
Principal Consultant
Ridgway, CO
United States
© Copyright 2023 by ComplianceAcuity, Inc. All rights reserved.
------------------------------
Original Message:
Sent: 27-Jan-2023 09:50
From: Anonymous Member
Subject: Design and Risk Assessment Requirements For Class I Device
This message was posted by a user wishing to remain anonymous
Our company has only historically produced FDA Class II devices which required clearance through a 510(k). We are now considering altering the intended use and design of one of our product lines into a Class I 510(k) exempt device. We have reviewed the general controls requirements but are a little uncertain which pieces of our Quality System that exists for our Class II devices applies to the new Class I device. Do we need all of the normal design input and output documents? All of the same risk documentation per ISO 14971 (dFMEA, PFMEA)? Also does the answer change if the device is exempt for the current good manufacturing practices?