Hello Ming,
I completely agree with your conclusion that a Clinical Evaluation Report (CER) is not required for the type of combination product being developed.
The basic requirement for drug delivery combination products is described in the second paragraph of MDR Article 1(9), which requires:
". . . the relevant general safety and performance requirements set out in Annex I to this Regulation shall apply as far as the safety and performance of the device part of the single integral product are concerned."However, under certain circumstances, there are additional special requirements that need to be met, and these are described in MDR Article 117, which offers two possibilities for meeting the regulatory requirements applicable to such drug delivery combination products.
The first paragraph of Article 117 addresses the special situation where the device part is a CE marked medical device in its own right. If this is the case, then this CE marked device must meet the entire MDR, including the requirement for a CER. This does not, however, appear to be the case for the combination product you are describing.
The second paragraph of Article 117 addresses the situation where the device part of a drug delivery combination product is not CE marked in its own right. It does, however, meet the applicable GSPRs for the device part being used in the combination product. Under this circumstance, which I believe is the situation described in your question,
no CER is required, because a CER is not included within the MDR GSPRs (as it was under the MDD).
In fact, in the second paragraph of Article 117, there are two routes to follow, depending on what would be the classification of the device part, were it to be assessed as a stand-alone medical device.
If the device part would fall under an MDR classification that required notified body involvement, the applicant (your company) would need to obtain a written opinion from a notified body designated for that type of device under the MDR to confirm its compliance with the applicable GSPRs and submit that opinion with the combination product application to the drug authority.
However, Article 117 of the MDR does not address the circumstance where the device part, if used separately, would be classified as Class I, so based on the language of the Article, a confirmation of compliance with the applicable GSPRs, such as a completed and signed GSPR Checklist, without the involvement of a notified body, should be sufficient to meet the requirement of MDR Article 1(9).
Finally, attention has been called by earlier responders to MDR Article 5.3 and Article 61; however, these MDR articles apply to an item that is regulated solely as a medical device and not as a medicinal product.
------------------------------
Maria E Donawa, M.D.
President, Donawa Lifescience
Rome, Italy
medonawa@donawa.com+39 06 578 2665
------------------------------
Original Message:
Sent: 22-Mar-2021 01:29
From: Ming Aquila
Subject: Question regarding MDR Article 117 and Clinical Evaluation Report
We are developing a combination product with the drug PMOA. The combination product meets the definition per MDR Article 1 Point 9 the second paragraph, therefore, we are following MDR Article 117 requirements. There are directions given by consultants that we need to produce a CER (clinical evaluation report). I wonder if this is indeed a requirement as Article 117 does not dictate such a requirement. Certainly, an executive summary must include that the device constituent part is deemed safe and effective as the combination product as a whole performs as intended; however, from my point of view, a CER is not warranted. as we are evaluating clinical efficacy from a drug (PMOA) perspective, not the device. Can someone who has experience dealing with NB Opinion on a Combination Product with drug PMOA shed some light on this topic? Thank you.